Default Are these radon results fubar? Updated

Taken with a sun nuclear model 1027.

Comments?

The data is read from left to right and each data point represents one hour of testing.

tape1.jpg

These were taken 6 minutes apart and the second tape was taken just before the last hour data point was recorded. Note the very different averages. I think the machine started to average the results in the electronics just before the last hour was integrated and recorded.

tape2.jpg

The graph immediately got my attention. I have not seen this shape of graph with such distinct spikes. When the homeowner was asked about laundry, the time frames that washing was done coincides with these spikes.

tape3.jpg

These results were printed within minutes of each other after hour #48. Note the very different averages and also note the last hour with a radon measurement of 500 pCi/L.

tape4.jpg

this shows the data points before and after hour #48. Note the different values recorded for the hours shown.

these-radon-results-fubar-tape1.jpg tape1.jpg

these-radon-results-fubar-tape2.jpg tape2.jpg

these-radon-results-fubar-tape3.jpg tape3.jpg

these-radon-results-fubar-tape4.jpg tape4.jpg

Has anyone else ever seen something like this?

Thanks, Bill

Last edited by wthacker; 2/24/09 at 2:56 PM..
Reply With Quote
  #2  
Old 2/23/09, 10:54 PM
Scott F. Coslett Scott F. Coslett is offline
InterNACHI Member
 
Join Date: Mar 2005
Location: Berwick, PA
Posts: 193
Default Re: Are these radon results fubar?

William -

Can you post associated graphs for tapes #3 and #4??



Scott Coslett
National Property Inspections
Berwick, PA
www.PennInspector.com
scoslett@pa.metrocast.net


"Life is Tough. Its even tougher if you're stupid" John Wayne
Reply With Quote
  #3  
Old 2/24/09, 7:10 AM
David C. Macy's Avatar
David C. Macy David C. Macy is offline
InterNACHI Member
 
Join Date: Jan 2004
Location: NE. OH
Posts: 2,657
Default Re: Are these radon results fubar?

That appears to be an invalid test. Minimum test time is 48 hours. I only see 47 hours recorded. In regards to the spikes. I use SUN monitors and have never had such discrepancies. Has the machine been calibrated properly? Have cross checks and duplicate tests been done?

I would not use those test results.
Reply With Quote
  #4  
Old 2/24/09, 7:59 AM
Frank J. Carey Frank J. Carey is offline
InterNACHI Member
 
Join Date: Feb 2008
Location: Exeter, PA
Posts: 113
Default Re: Are these radon results fubar?

get your monitors checked



Frank

Carey Inspection Solutions, LLC
Wilkes-Barre/ Scranton PA
Nachi ID -> 08022204
Frank@careyinspectpa.com
www.careyinspectpa.com
www.frankcareyconstruction.com
Reply With Quote
  #5  
Old 2/24/09, 8:50 AM
William Thacker's Avatar
William Thacker William Thacker is offline
InterNACHI Member
 
Join Date: Dec 2008
Location: Pennsylvania
Posts: 354
Default Re: Are these radon results fubar?

Quote:
Originally Posted by dmacy View Post
That appears to be an invalid test. Minimum test time is 48 hours. I only see 47 hours recorded. In regards to the spikes. I use SUN monitors and have never had such discrepancies. Has the machine been calibrated properly? Have cross checks and duplicate tests been done?

I would not use those test results.
Thanks David,

Tape 3 shows the results before and after hour #48. The first tape was printed just a few minutes before the last hour was of data was recorded. This test was performed using the same machine. It was calibrated last July and had a cross check performed last month. It was spot on.

I don't look at the spikes as a discrepancy. (except for perhaps hour #48 with a level of 500 pCi/L. I think it is related to doing laundry. The laundry room was adjacent to the basement club room where the detector was located. The gas water heater and furnace are also in the laundry room.

The home has an active mitigation system, installed less than 2 years ago. It was pulling 3/4" of water. The post mitigation test was less than 2.0 pCi/L but it was performed prior to the owners occupying the home.

I did not certify these results. I commenced a retest with a different CRM. I am picking that monitor up in a few hours and will update you with those results.

Bill
Reply With Quote
  #6  
Old 2/24/09, 8:53 AM
William Thacker's Avatar
William Thacker William Thacker is offline
InterNACHI Member
 
Join Date: Dec 2008
Location: Pennsylvania
Posts: 354
Default Re: Are these radon results fubar?

Quote:
Originally Posted by fcarey View Post
get your monitors checked
I think I'll be sending it to the manufacturer for repairs and recalibration.

Thanks Frank,

Bill
Reply With Quote
  #7  
Old 2/24/09, 8:54 AM
William Thacker's Avatar
William Thacker William Thacker is offline
InterNACHI Member
 
Join Date: Dec 2008
Location: Pennsylvania
Posts: 354
Default Re: Are these radon results fubar?

Quote:
Originally Posted by scoslett View Post
William -

Can you post associated graphs for tapes #3 and #4??
Here is the graph for the tape3.jpg Scott.

For tape4.jpg, this is the graph for the right side. tape2.jpg, posted above, is the graph for the left side.

Bill
Attached Thumbnails
these-radon-results-fubar-tape5.jpg  

Last edited by wthacker; 2/24/09 at 2:36 PM.. Reason: added graph
Reply With Quote
  #8  
Old 2/24/09, 2:46 PM
William Thacker's Avatar
William Thacker William Thacker is offline
InterNACHI Member
 
Join Date: Dec 2008
Location: Pennsylvania
Posts: 354
Default Re: Are these radon results fubar?

The retest performed with a different detector shows the expected results which are very much different than the previous 48 hours.

these-radon-results-fubar-tape7a.jpg (white out for client confidentiality)

these-radon-results-fubar-tape8.jpg

I contacted my radon guru and his assessment is as follows:

"If you can rule out customer interference, and instrument noise, then you may have an extremely tender house. By that I mean a house were small changes in differential pressure greatly affects radon transport."

"If the same machine for both tapes, and if the weather could not be the problem for the times indicated, you may have a bad machine."

Frank was looking for a used 1027......

"A reading of 500 pCi/L would require water radon levels of probably more than 5 million pCi/L of water---- I don't think so."

"Also, I don't think the radon water concentration would vary that much. As you know, the weather was very cold with high winds."

"The last reading should not have changed the other hourly readings. This is not true for the averages. You can now see why I always expose a charcoal canister along with EVERY machine test."

"I would repeat the test with duplicated detectors."


It is easy for him to use charcoal canisters. He manufacturers and analyzes them.

Last edited by wthacker; 2/24/09 at 2:58 PM..
Reply With Quote
  #9  
Old 2/25/09, 7:04 AM
Rick K. Kie's Avatar
Rick K. Kie Rick K. Kie is offline
InterNACHI Member
 
Join Date: Jan 2008
Location: Endwell, NY
Posts: 385
Default Re: Are these radon results fubar?

The last reading said 500 not 500.00 it looks fishy to me. I have never ran into this either. Also notice it wasn't averaged in the test the first and last results are tossed if I am not mistaken to get the EPA average. Although it should have been addded into the first average. Maybe the unit was unplugged from the computer to early or there was a glitch.

I would use a different monitor and chect it again or better yet use two 1027's to retest.



Rick Kie

Last edited by rkie; 2/25/09 at 7:07 AM..
Reply With Quote
  #10  
Old 2/25/09, 8:13 AM
Frank J. Carey Frank J. Carey is offline
InterNACHI Member
 
Join Date: Feb 2008
Location: Exeter, PA
Posts: 113
Default Re: Are these radon results fubar?

Hahaha don't get any ideas William. I am interested to see what you come up with with this.



Frank

Carey Inspection Solutions, LLC
Wilkes-Barre/ Scranton PA
Nachi ID -> 08022204
Frank@careyinspectpa.com
www.careyinspectpa.com
www.frankcareyconstruction.com
Reply With Quote
  #11  
Old 2/25/09, 5:14 PM
cduphily's Avatar
cduphily cduphily is offline
Active Poster
 
Join Date: Sep 2006
Location: Dingmans Ferry, PA
Posts: 2,486
Please Note: cduphily is a non-member guest and is in no way affiliated with InterNACHI or its members.
Default Re: Are these radon results fubar?

Quote:
Originally Posted by dmacy View Post
That appears to be an invalid test. Minimum test time is 48 hours. I only see 47 hours recorded. In regards to the spikes. I use SUN monitors and have never had such discrepancies. Has the machine been calibrated properly? Have cross checks and duplicate tests been done?

I would not use those test results.
I'll have to find the exact language but PA - DEP gives 2-3 hours leeway between placement/pickup for ease of scheduling. I typically leave my monitors an extra 10-12 hrs longer if I can not pick up close to the drop off time (rural area) ... but some CAN pick up sooner.
Reply With Quote
  #12  
Old 2/25/09, 9:54 PM
William Thacker's Avatar
William Thacker William Thacker is offline
InterNACHI Member
 
Join Date: Dec 2008
Location: Pennsylvania
Posts: 354
Default Re: Are these radon results fubar?

Quote:
Originally Posted by cduphily View Post
I'll have to find the exact language but PA - DEP gives 2-3 hours leeway between placement/pickup for ease of scheduling. I typically leave my monitors an extra 10-12 hrs longer if I can not pick up close to the drop off time (rural area) ... but some CAN pick up sooner.
Hi Chris,

For a 48 hour test, you have 48 + 2 hours to pick it up. After that for each additional 24 hours, you have plus or minus 2 hours.

Picking up a detector 58 to 60 hours later is not protocol. Do you certify a 60 hour test?

Bill
Reply With Quote
  #13  
Old 2/25/09, 9:56 PM
William Thacker's Avatar
William Thacker William Thacker is offline
InterNACHI Member
 
Join Date: Dec 2008
Location: Pennsylvania
Posts: 354
Default Re: Are these radon results fubar?

Quote:
Originally Posted by fcarey View Post
Hahaha don't get any ideas William. I am interested to see what you come up with with this.
Dang it....

I am conducting a co-located test right now with the suspect CRM and another CRM. So far they are in good agreement.
Reply With Quote
  #14  
Old 2/25/09, 9:57 PM
William Thacker's Avatar
William Thacker William Thacker is offline
InterNACHI Member
 
Join Date: Dec 2008
Location: Pennsylvania
Posts: 354
Default Re: Are these radon results fubar?

Quote:
Originally Posted by rkie View Post
The last reading said 500 not 500.00 it looks fishy to me. I have never ran into this either. Also notice it wasn't averaged in the test the first and last results are tossed if I am not mistaken to get the EPA average. Although it should have been addded into the first average. Maybe the unit was unplugged from the computer to early or there was a glitch.
The unit drops the decimal once radon levels go triple digit. The results were puzzling indeed though Rick.
Reply With Quote
  #15  
Old 2/26/09, 7:34 AM
Frank J. Carey Frank J. Carey is offline
InterNACHI Member
 
Join Date: Feb 2008
Location: Exeter, PA
Posts: 113
Default Re: Are these radon results fubar?

I will also from time to time leave a monitor for an extra 12 if I am not sure closed conditions are satisfied when I get there. These monitors only take the last 48 measurements under EPA avg. A short term test is defined as 2-90 days I believe. Don't quote me on that.



Frank

Carey Inspection Solutions, LLC
Wilkes-Barre/ Scranton PA
Nachi ID -> 08022204
Frank@careyinspectpa.com
www.careyinspectpa.com
www.frankcareyconstruction.com
Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are On
Pingbacks are On
Refbacks are On


Similar Threads
Thread Thread Starter Forum Replies Last Post
Documents on which InterNACHI's radon course content is based: gromicko IAC2 Indoor Air Forum 4 1/18/09 6:53 AM
Radon Testing Kansas City dmoriconi Ancillary Inspection Services & Additional Topics 1 1/11/09 3:11 PM
Final review of Newman's radon references! Caoimhín P. Connell Ancillary Inspection Services & Additional Topics 12 10/4/08 12:23 PM
Radon Class in Oakville and Ottawa mkyriacou Canadian Inspectors 9 2/2/08 11:48 PM


All times are GMT -4. The time now is 10:14 AM.


Popular Sections

:

All Sections

Inspection News

InterNACHI Membership

Inspection Standards

Inspection Education

InterNACHI Inspectors

Inspection Links

 

 

 

NACHI.ORG Statistics

 

 

no new posts