I have heard description of this unit that describes many times the image freezes on the screen and it retains a shadow even if the camera was sweeping at 30 degree angles non stop.

Is this a normal occurence or is the unit defective? I’m particularly concerned because I haven’t seen a FLIR i7 before and wanted to buy one for general use and I don’t want it to pause every now and then.

Is the reason for the pause because the temperature scale is adjusting? Does it do this every few degrees of turns?

What kind of thermal imager that don't do this constant pausing? I don't think it has to do with 9Hz because the pause is a second long or so. Thanks.
Reply With Quote
Need a home inspection in Illinois? Check out InterNACHI's listing of Illinois certified home inspectors. Or, find a home inspector anywhere in the world with our inspection search engine.
  #2  
Old 8/11/11, 9:33 AM
David A. Andersen's Avatar
David A. Andersen David A. Andersen is offline
InterNACHI Member
 
Join Date: Jun 2005
Location: Woodlawn, TN
Posts: 5,927
Default Re: FLIR i7 question

All cameras do that. Some more than others.

This is to correct for drift.
If the camera does not do it, it will drift right off the scale.

Turn the camera on for several minutes before you want to use it and don't shut it off till your done. Also leave the lens cap open.

Adverse temps indoors/out will make this happen more.
Bring in the camera and leave it out of the case till you need it.



"Insanity: doing the same thing over and over again and expecting a different results." Albert Einstein

David A. Andersen & Associates
Clarksville - Nashville Home Inspector Lic#40
http://www.midtninspections.com
ITC Level III Thermographer Cert#1958
Building Science Thermographer Cert#33784
http://www.thermalimagingscan.com
HVAC Certification EPA Cert#2046620
BPI# 5015804
Link to my Website at: http://www.midtninspections.com/link-submission
Reply With Quote
  #3  
Old 8/11/11, 10:12 AM
Rye Zers Rye Zers is offline
New User
 
Join Date: Aug 2011
Posts: 28
Please Note: Rye Zers is a non-member guest and is in no way affiliated with InterNACHI or its members.
Default Re: FLIR i7 question

Quote:
Originally Posted by dandersen View Post
All cameras do that. Some more than others.

This is to correct for drift.
If the camera does not do it, it will drift right off the scale.
What do you mean "it will drift right off the scale"? I think it's auto adjusting the temperature range and unit with faster processor or graphics can re adjust faster, right? This means if I "lock" the screen or temperature, it won't pause anymore.

Quote:
Turn the camera on for several minutes before you want to use it and don't shut it off till your done. Also leave the lens cap open.

Adverse temps indoors/out will make this happen more.
Bring in the camera and leave it out of the case till you need it.
Besides inspecting electrical. I want to use it too at night in quick scan of the outside of my house to scan for animals hiding in bushes. And I don't want it to pause while scanning. Maybe by "locking" the tempeture, it will avoid the pause and temperature rescaling?
Reply With Quote
  #4  
Old 8/11/11, 11:31 AM
Linas Dapkus's Avatar
Linas Dapkus Linas Dapkus is offline
InterNACHI Member
 
Join Date: Feb 2011
Location: Home
Posts: 2,187
Default Re: FLIR i7 question

Quote:
Originally Posted by dandersen View Post
All cameras do that. Some more than others.

This is to correct for drift.
If the camera does not do it, it will drift right off the scale.

Turn the camera on for several minutes before you want to use it and don't shut it off till your done. Also leave the lens cap open.

Adverse temps indoors/out will make this happen more.
Bring in the camera and leave it out of the case till you need it.
Do you own an I-7 Dave? I heard that sure is a mighty fine piece of equipment.
Reply With Quote
  #5  
Old 8/11/11, 3:25 PM
David A. Andersen's Avatar
David A. Andersen David A. Andersen is offline
InterNACHI Member
 
Join Date: Jun 2005
Location: Woodlawn, TN
Posts: 5,927
Default Re: FLIR i7 question

I'm sorry Sir, I have never held one and I am just doing the best I can with the information provided...

Quote:
What do you mean "it will drift right off the scale"?
Rye; to understand what I am talking about may take some continuing education that I am not prepaired to attempt to convey in a short message board post.

If you do not understand "Drift" I recommend that you take the required courses before you pruchase or use any thermal imaging device so you don't wast money and go off "half-cocked" on some adventure...

Or just ask John, he can square you away.



"Insanity: doing the same thing over and over again and expecting a different results." Albert Einstein

David A. Andersen & Associates
Clarksville - Nashville Home Inspector Lic#40
http://www.midtninspections.com
ITC Level III Thermographer Cert#1958
Building Science Thermographer Cert#33784
http://www.thermalimagingscan.com
HVAC Certification EPA Cert#2046620
BPI# 5015804
Link to my Website at: http://www.midtninspections.com/link-submission
Reply With Quote
  #6  
Old 8/11/11, 6:33 PM
Rye Zers Rye Zers is offline
New User
 
Join Date: Aug 2011
Posts: 28
Please Note: Rye Zers is a non-member guest and is in no way affiliated with InterNACHI or its members.
Default Re: FLIR i7 question


I know all about blackbody radiation and emissivity. Anyway. Thermal drift is defined as "Drift caused by internal heating of equipment during normal operation or by changes in external ambient temperature".

With regards to the screen pausing every often. I think what its doing is auto adjusting the temperature scale depending on target. For example. When you point the imager at a scene with temperature between 15 to 50 Celsius. The screen will show that scale. If you then point it to higher temperature. It will pause as the screen updates and turns the scale to 15 to 120 Celsius. I think this is the reason for the screen pause. I wonder if the unit microprocessor speed is related to how fast it updates. To those who own the FLIR i7. How does it perform compare to others? I plan to order it as soon as I get this info. Thanks.
Reply With Quote
  #7  
Old 8/11/11, 6:46 PM
Chuck Evans's Avatar
Chuck Evans Chuck Evans is offline
InterNACHI Member
 
Join Date: Sep 2006
Location: Stagecoach, TX
Posts: 2,037
Default Re: FLIR i7 question

Research NUC or Non-Uniformity Correction and trust Dave's word that it is a normal and essential function of the imager.



Chuck Evans (TREC #7657)
Level III Infraspection Institute Certified Infrared Thermographer (#8402)
HomeCert Houston Home Inspections & Thermal Inspections Find us on Facebook
Houston Thermal Inspections & Infrared Imaging Find us on Facebook
Houston Home Inspector

Houston, TX
Reply With Quote
  #8  
Old 8/12/11, 4:34 AM
Rye Zers Rye Zers is offline
New User
 
Join Date: Aug 2011
Posts: 28
Please Note: Rye Zers is a non-member guest and is in no way affiliated with InterNACHI or its members.
Default Re: FLIR i7 question

I went to a FLIR i7 demo center and went to work with it. Temperature scale changes don't make the screen pause. So it is the Non-Uniformity Correction that causing the pauses. Now for those who own higher end models. Is there a way to temporary disable the feature in situation that you need fast motion and can take a slight decrease in temperature accuracy (such as mobile based thermal imager like the PathFinder car system or military vehicle thermal system)? Just curious as I know the FLIR i7 won't make you disable it.
Reply With Quote
  #9  
Old 8/12/11, 12:35 PM
Scott Gilligan,  CMI's Avatar
Scott Gilligan, CMI Scott Gilligan,  CMI is offline
InterNACHI Member
 
Join Date: May 2007
Location: Philadelphia, PA
Posts: 1,828
Send a message via AIM to sgilligan1 Send a message via MSN to sgilligan1 Send a message via Yahoo to sgilligan1
Default Re: FLIR i7 question

First off...you are not going to get any kind of advanced features like that from a low end camera. If you want live video that does not get interrupted, that is not something that modern handheld cameras can do because they are uncooled microbolometers. You would need to think about going into the cooled focal plane array camera world or into a surveillance system type camera.

No one camera is going to be able to perform all tasks, especially lower end entry level cameras. You are basically trying to make a hammer perform as a calculator.

Pick which application you want to do and find the appropriate camera to do the job.
Reply With Quote
  #10  
Old 8/12/11, 1:57 PM
John McKenna's Avatar
John McKenna John McKenna is offline
InterNACHI Member
 
Join Date: Jun 2006
Location: Crockett, Tx
Posts: 12,303
Default Re: FLIR i7 question

Quote:
Originally Posted by sgilligan1 View Post
First off...you are not going to get any kind of advanced features like that from a low end camera. If you want live video that does not get interrupted, that is not something that modern handheld cameras can do because they are uncooled microbolometers. You would need to think about going into the cooled focal plane array camera world or into a surveillance system type camera.

No one camera is going to be able to perform all tasks, especially lower end entry level cameras. You are basically trying to make a hammer perform as a calculator.

Pick which application you want to do and find the appropriate camera to do the job.
Exactly.



John McKenna, CMI (TREC #4565)
Executive Director - Master Inspector Certification Board
25 Yrs Constr Exp - 13 Yrs Home Inspector Exp
American Home Inspection - East Texas.

Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are On
Pingbacks are On
Refbacks are On


Similar Threads
Thread Thread Starter Forum Replies Last Post
Flir B50 errors and problems jmeeker Thermal Imaging, Infrared Cameras & Energy Audits 9 7/19/11 11:30 PM
Sump Pit/Pump Question Joey D'Adamo Plumbing Inspections 4 11/26/08 1:04 PM
Is FLIR infared imaging the future wave of Home Inspections? jwatts1 General Inspection Discussion 66 5/1/07 5:48 PM


All times are GMT -4. The time now is 7:28 PM.


Popular Sections

:

All Sections

Inspection News

InterNACHI Membership

Inspection Standards

Inspection Education

InterNACHI Inspectors

Inspection Links

 

 

 

NACHI.ORG Statistics

 

 

no new posts