  #16  
Old 8/19/11, 2:22 PM
Jason Kaylor Jason Kaylor is offline
InterNACHI Member
 
Join Date: Feb 2009
Location: Tempe, AZ
Posts: 970
Send a message via Yahoo to jkaylor
Default Re: thermal sensitivity vs thermal accuracy

Rye,

That depends if you consider that the +/- 2% will put you at 96.63F to 100.57F of human body temperature. 100.57F would be considered a fever for a person that is running at 98.6F.

Personally, I think all the border IR and fever IR is something to make the masses feel good about "things being taken care of".

The others here are correct. At the very low end of the spectrum you should buy at least a FLIR E40/E40BX. If you can afford it, go up to the E50/E60. They are also correct in the fact that training is more valuable in the long run. Without the ability to adjust focus the I7 is not really considered any type of professional IR camera. They are really good for field guys of electrical companies or power companies that want to do some quick shots of panels, etc (just as an example)

JJ
Reply With Quote
Need a home inspection in Montana? Check out InterNACHI's listing of Montana certified home inspectors. Or, find a home inspector anywhere in the world with our inspection search engine.
  #17  
Old 8/19/11, 4:23 PM
David A. Andersen's Avatar
David A. Andersen David A. Andersen is offline
InterNACHI Member
 
Join Date: Jun 2005
Location: Woodlawn, TN
Posts: 5,927
Default Re: thermal sensitivity vs thermal accuracy

As posted, you must decide the application and pick the camera.

There is nothing wrong with the I7 (for the intended purpose).

You are so all over the board on what you want to use and camera specs from $2k - $89k range, I am lost...

You can not do electrical/mechanical inspections with an I7 and go to the airport and prevent SARS Virus from entering your country...



"Insanity: doing the same thing over and over again and expecting a different results." Albert Einstein

David A. Andersen & Associates
Clarksville - Nashville Home Inspector Lic#40
http://www.midtninspections.com
ITC Level III Thermographer Cert#1958
Building Science Thermographer Cert#33784
http://www.thermalimagingscan.com
HVAC Certification EPA Cert#2046620
BPI# 5015804
Link to my Website at: http://www.midtninspections.com/link-submission
Reply With Quote
  #18  
Old 8/19/11, 4:49 PM
Jason Kaylor Jason Kaylor is offline
InterNACHI Member
 
Join Date: Feb 2009
Location: Tempe, AZ
Posts: 970
Send a message via Yahoo to jkaylor
Default Re: thermal sensitivity vs thermal accuracy

Quote:
Originally Posted by dandersen View Post
go to the airport and prevent SARS Virus from entering your country...
How about Ebola?

JJ
Reply With Quote
  #19  
Old 8/19/11, 6:47 PM
Linas Dapkus's Avatar
Linas Dapkus Linas Dapkus is offline
InterNACHI Member
 
Join Date: Feb 2011
Location: Home
Posts: 2,187
Default Re: thermal sensitivity vs thermal accuracy

Quote:
Originally Posted by jkaylor View Post
How about Ebola?

JJ
They just came out with a fart detection thermal imager. It has video capability too!!
Reply With Quote
  #20  
Old 8/19/11, 6:47 PM
Rye Zers Rye Zers is offline
New User
 
Join Date: Aug 2011
Posts: 28
Please Note: Rye Zers is a non-member guest and is in no way affiliated with InterNACHI or its members.
Default Re: thermal sensitivity vs thermal accuracy

Quote:
Originally Posted by jkaylor View Post
Rye,

That depends if you consider that the +/- 2% will put you at 96.63F to 100.57F of human body temperature. 100.57F would be considered a fever for a person that is running at 98.6F.

Personally, I think all the border IR and fever IR is something to make the masses feel good about "things being taken care of".

The others here are correct. At the very low end of the spectrum you should buy at least a FLIR E40/E40BX. If you can afford it, go up to the E50/E60. They are also correct in the fact that training is more valuable in the long run. Without the ability to adjust focus the I7 is not really considered any type of professional IR camera. They are really good for field guys of electrical companies or power companies that want to do some quick shots of panels, etc (just as an example)

JJ
Too bad I only hear this now that I have ordered the I7 from Estonia (all I7 are produced in Estonia). But I don't like manual focus. I want to also use the i7 at streets or close parking to detect which cars are recently parked (heatness of the engine and hood) and I don't want to keep focusing lens as it can lose time and make people notice. At home and office. I'd use the i7 to check on electrical panels to see which is breaker hotter and need replacement. Also to check on whether certain outlets and wires have enough thickness to be resistant to dangerous heat. Also inspecting what part of a room can cause heat making aircon lost efficiency. i7 is enough isn't it. If not why not so I can maybe cancel my order at Estonia.

For me. The i7 is more of home use as well as field use. Admit it that a chechen rebel would benefit from even the i7. When in dark neightborhood. I want to use the i7 to quickly see if there are hidden people in the vicinity. And I don't want to use manual focus and adjusting it slowly because in tactical situations. Time is of the essence.
Reply With Quote
  #21  
Old 8/19/11, 8:57 PM
Chuck Evans's Avatar
Chuck Evans Chuck Evans is offline
InterNACHI Member
 
Join Date: Sep 2006
Location: Stagecoach, TX
Posts: 2,037
Default Re: thermal sensitivity vs thermal accuracy

I think it's time to banish this clown



Chuck Evans (TREC #7657)
Level III Infraspection Institute Certified Infrared Thermographer (#8402)
HomeCert Houston Home Inspections & Thermal Inspections Find us on Facebook
Houston Thermal Inspections & Infrared Imaging Find us on Facebook
Houston Home Inspector

Houston, TX

Last edited by cevans; 8/20/11 at 5:42 PM..
Reply With Quote
  #22  
Old 8/20/11, 9:54 AM
Rye Zers Rye Zers is offline
New User
 
Join Date: Aug 2011
Posts: 28
Please Note: Rye Zers is a non-member guest and is in no way affiliated with InterNACHI or its members.
Default Re: thermal sensitivity vs thermal accuracy

I know the FLIR i7 is a severe threat to home inspection business because home owners can buy their own thermal imager and check where the aircon is leaking. But don't worry.... because companies can be inspected only using your expertise so why not try companies inspection instead of just home. This way. the i7 is not a threat to your business.

I ordered the i7 the other day because the next cheapest 160x120 is the e40 and it costs $4500 in the US. At least the i7 at $1995 is affordable enouigh.
Reply With Quote
  #23  
Old 8/21/11, 5:18 PM
John Gromkoski's Avatar
John Gromkoski John Gromkoski is offline
InterNACHI Member
 
Join Date: Apr 2007
Location: Staten Island, NY
Posts: 608
Default Re: thermal sensitivity vs thermal accuracy

Quote:
Originally Posted by cevans View Post
I think it's time to banish this clown

I couldn't agree more.




774 Manor Road
Staten Island, NY 10314
718-514-3393
Reply With Quote
  #24  
Old 8/21/11, 7:25 PM
David A. Andersen's Avatar
David A. Andersen David A. Andersen is offline
InterNACHI Member
 
Join Date: Jun 2005
Location: Woodlawn, TN
Posts: 5,927
Default Re: thermal sensitivity vs thermal accuracy

So take your $1995 camera and use it like all the homeowners that have purchased them and found out that it is not a "point and shoot" camera.

Go to the ITC (Infrared Training Center) web site and look at all the nice vidieos that people spent many hours producing to answer questions just like yours.

There is no one here that is qualified and has the time to teach you.



"Insanity: doing the same thing over and over again and expecting a different results." Albert Einstein

David A. Andersen & Associates
Clarksville - Nashville Home Inspector Lic#40
http://www.midtninspections.com
ITC Level III Thermographer Cert#1958
Building Science Thermographer Cert#33784
http://www.thermalimagingscan.com
HVAC Certification EPA Cert#2046620
BPI# 5015804
Link to my Website at: http://www.midtninspections.com/link-submission
Reply With Quote
  #25  
Old 8/23/11, 1:27 PM
Rye Zers Rye Zers is offline
New User
 
Join Date: Aug 2011
Posts: 28
Please Note: Rye Zers is a non-member guest and is in no way affiliated with InterNACHI or its members.
Default Re: thermal sensitivity vs thermal accuracy

Quote:
Originally Posted by cevans View Post
Thermal sensitivity and thermal accuracy are different specifications. There is another factor that you need to be aware of which is spot size. An infrared imager (imaging radiometer) cannot accurately measure temperatures at the single sensor (pixel level)

Spot size defines the minimum size object that can be accurately measured with a given lens at a particular distance with a particular imager. Anything smaller than the spot size cannot be accurately measured, although it may be readily seen. The spot size will cover some number of sensors and is an important consideration when performing quantitative anaysis.

Thermal sensitivity specifies the minimum change in signal due to variation in incident radiation that the sensor can detect (i.e., how small a temperature variation can the imager see). The sensor can detect temperature variations in things that are too small for it to measure.

Each sensor in the array will have a different response curve to infrared radiation. Each cell in the array will vary in signal response from other cells around it for the same incident radiation. This is corrected through the camera firmware (set during the factory calibration process) and is periodically normalized through the NUC (non-uniformity correction) process, which corrects for drift. Most current imagers NUC automatically.

Basically, your imager will periodically flip a uniform temperature screen up in front of the sensor and readjust for the signal level of each pixel in the array so that it renders the same temperature across the whole array. Think of it like the mirror in an SLR camera. Every time your imager freezes and makes the familiar clicking noise, which is the screen flipping up in front of the sensor, it is going through the NUC process.

The NUC process is what eliminates the "smearing" that you refer to.

This is also the reason that your uncooled imager should be given time to warm-up and reach equilibrium in the environment in which you are using it. You may also notice improved image clarity an accuracy immediately after the NUC event.

This is but one of many reasons that thermography is not simply point and shoot. If you are doing energy audits you should be focused on the thermal patterns (qualitative analysis)

Yes you can see significant differences between imagers with different spacial resolutions and thermal sensitivities.
Very excellent description. But for those that says speciificaly that 0.05 c at 30 degree. Does this means at 50 Celius or 80 Celsius. The 0.05 C sensitivity no longer works? I thought from the begnning hte 0.05C sensitivity is across all temperature range. If not true. Then at 100 Celius, what is the thermal sensitivity.. would be become like 3 Celsius or something?
Reply With Quote
  #26  
Old 8/25/11, 2:43 AM
Rye Zers Rye Zers is offline
New User
 
Join Date: Aug 2011
Posts: 28
Please Note: Rye Zers is a non-member guest and is in no way affiliated with InterNACHI or its members.
Default Re: thermal sensitivity vs thermal accuracy

Let me just share what I found out about the theoretical limit or maximum megapixel for thermal camera before I leave this forum and join an actual IR class and give you guys peace

The formula for airy disc diameter is 2.44 * wavelength * focal-ratio.
For visible light wavelength of 550 nanometer (green light) and focal-ratio of 1.5 (most thermal imagers have focal ratio less than 2 to make the spot smaller as larger focal ratio like f/10 can make it larger), airy disc diameter is 2.44 * 550 nanometer * 1.5 =
2 micron. This is for visible light camera.

For thermal imagers. Wavelength is say 10 micron (middle of 7-13 micron) and focal ratio of 1.5 gives 2.44 * 10 micron * 1.5 = 36.6 micron

That is. Thermal camera spot size is 18 times larger than that of visible light therefore there is an upper limit to the megapixels.

From 36.6 micron... signal sample theorem gives about 18 micron minimum size pixel. Anything smaller and it's a waste of megapixel (it's not like in telescope where you can use barlows to increase pixel scale or make smaller pixel to image better double stars).

Now if you will check out the specs of FLIR cameras. They use about 17 micron for the SC600 for 640x480 and the camera is much larger than the E60 or Fluke Ti32 which only use 320x240.

Bottom line is. Because of the 36 micron spot diameter of the airy discs of thermal cameras. 320x240 or slighter larger camera for 640x480 is the ultimate megapixel as far as hand held portable thermal imager is concerned. They have now the 1 MP Atom thermal camera but it is quite large and not unlike the E60 or SC6000. Of course one can make 2Mp or even 10 MP thermal imagers but the camera would be as big as a desktop computer or even room size. This is because you have to create very large lens as the chip of 2 MP thermal camera that uses minimum 17 micron is larger.

Conclusion. 320x240 and 640x480 thermal camera in *portable* handheld format would be here to say for the next hundred years. And therefore it is not a bad idea to invest in a 640x480 as it will be useful for generations to come where even your grandchildren can still make good use of it.
Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are On
Pingbacks are On
Refbacks are On


Similar Threads
Thread Thread Starter Forum Replies Last Post
Mold Inspectors in Coeur d’Alene, ID 83815 rspriggs Ancillary Inspection Services & Additional Topics 15 1/3/12 1:14 PM
How about it illinois cbottger Thermal Imaging, Infrared Cameras & Energy Audits 13 8/19/11 6:07 PM
Look out John M here I come cbottger Thermal Imaging, Infrared Cameras & Energy Audits 14 12/28/08 11:01 AM
Ohio I know we can do it cbottger Thermal Imaging, Infrared Cameras & Energy Audits 0 12/26/08 12:31 AM


All times are GMT -4. The time now is 6:32 PM.


Popular Sections

:

All Sections

Inspection News

InterNACHI Membership

Inspection Standards

Inspection Education

InterNACHI Inspectors

Inspection Links

 

 

 

NACHI.ORG Statistics

 

 

no new posts